Description of the problem

Why and when should the 100k resistor between IN- (Pin7) and GND (Pin4) be used?


Description of the solution

This is to avoid the effects of common-mode voltage and it also works as a high impedance ground connection in case your signals are floating.


The Dewe-43 device is only Ch-GND isolated (so ISO-POWER), this is why the resistor is needed.

If non-isolated sensors would be connected to separate channels without the resistor between GND and IN- (4 to 7) it could cause different potentials on the GND connection and this could damage the device.



When the measurement is in single-ended mode connection type. Better than 100R resistor is a short between the In- and GND as shown in the images from the manual.



Example:

If the single-ended measurement is used, the Sirius will do a short between the GND and IN- which is the same as shown on the picture.

The DEWE-43 doesn't have this feature, which is why the BNC adapters were made with 100R resistor (IN- -> GND).


From our experience, the measurements have shown better if there is a lower resistance between IN- and GND regarding the noise and interference. 100kR connection is high impedance which can cause interference on the measurement and can cause more noise.


Conclusion:

It depends on what is more important for the customer. 

  1. If a high common mode voltage is expected --> then it is better to connect a resistor (larger resistance). 
  2. If a high noise is expected --> then it is better to connect the resistor directly to ground, or a small resistor, so that the grounds don't float, the input is protected, and there is less noise due to the lower impedance.


Additional information